View Single Post
DonDavis DonDavis is offline
What?!?!
 
DonDavis's Avatar
 
Join Date: Mar 2002
Location: Could be either Blountville Tenn or Mesa Az
Posts: 4,541
Garage
Installing an MRI, we had the magnet ramped and I was getting set up to perform SuperCon Shimming. That process makes the magnetic field Homogeneous which essentially means "make the field more smooth" by manipulating the several coil currents.

I was on top and began inserting the shim lead down into the vessel. I did not realize the vessel pressure had gone just a smidge below atmosphere and it pulled room temp air into the mix.

Heard an ever-so-slight air woosh and then the burst disk popped. Helium does not like being liquid and tries to be gas whenever it can. That exchange rate is roughly 800-1 meaning 1 liter of liquid becomes 800 liters of gas within seconds. I was sitting on about 2,000 liters of liquid.

**side note*** As soon as a magnet is physically put in place, we install the venting which vents to the outside. It's 12" diameter stainless steel the entire path.

The magnet quenched. "Oh s***" and the system began rumbling as the helium exited to outside. I scrambled down and met responders at the door stopping them from rushing in.

That one really sucked, but the convo with management was really great. A crucial step was not performed the day prior. Not my fault.

I still felt crappy, tho.

If that had been billable, it would have topped $100k without blinking.
__________________
running shoes, couple tools, fishing pole
1996 Subaru Legacy Outback AWD, 5speed
2002 Subaru Impreza WRX, 5speed
2014 Tundra SR5, 4x4
1964 Land Rover SII A 109 - sold this albatross
Old 09-27-2023, 03:08 PM
  Pelican Parts Catalog | Tech Articles | Promos & Specials    Reply With Quote #24 (permalink)